1. Auto removal of false data like spikes or removal of solvent peaks upon a given command. 2. Auto analysis about data “runaways” like false integrated peak areas because of signals are outside the linear or calibrated detector working range. 3. Removal of peak signals with excessive broadness but giving an ALERT. 4. Correction of retention time in case of temperature (GC) or flow (HPLC) problems during the analysis or for other purposes like precise chromatogram comparison. 5. Calculation of the retention index for all peaks (GC isothermal, temperature or pressure programmed and integration of the index values into the EXPORT chromatogram file). 6. Calculation of the k-values or, if phase programmed, of LC-index values and integration of the k-, or index values into the EXPORT chromatogram file. 7. Correction of selected peak areas by correction factors for quantitation purposes. 8. Calculation of specifically selected data like the n-alkane content in a hydrocarbon mix. 9. A new fully quantitative and well organized qualitative graphics display of the chromatogram allowing immediate understanding if two runs are compared or one run is seen relative to a standard chromatogram. This new look displays “line chromatograms”.
Many more other chromatogram “manipulation” is possible like seen in a command file field as given below but here is a strict statement:
The original raw chromatogram data file (binary) is NOT in any way touched, it remains as it was. The EXPORT data file however is a “reduced raw data” file as result of an integration run. Its correction, adaptation or enlargement is REVERSIBLE.
The integration program extracts from the binary RAW data file with thousands of signal data (intensity over time) the following values:
a) the retention time in seconds, found by 5th polynomial interpolation of seven raw data points around the peak top. This allows to take retention time data by +- 0.001 sec through interpolation. A quartz clock driven signal converter storing raw data however is necessary, otherwise high time precision is not available.
b) the peak width in half height in seconds, found automatically by mathematical statistical approaches and NOT based on the (false) assumption, the peak has a Gaussian shape. It has not: the variance of real peaks differs from the variance of a real GAUSS function easily by 20 to 40 and more percent even in capillary gas and capillary liquid chromatography. The peak width data are found by +- 0.001 sec. precision but see under a) above.
c) the peak height in amperes or volt (nano-, micro-, milli-... depending on the electronics)
d) the integrated peak area based on the best possible baseline positioning, given in volt x seconds or in amperes x seconds.
Into this reduced raw data file there is one column integrated, which contains zeros, or already some retention index values, or k-data or substance specific codes. At the end of the EXPORT data set there are two 255 positions long ASCII lines attached containing 12 data fields. Each has 20 characters capacity and is enclosed in “semicolon” delimiters. This line portions of 20 characters capacity each contain information like - year / month / day / hour / minute / second of sample injection; - institution / laboratory / instrument; - sample code taken where, when, by whom, how, after which treating and sampling mode; - analytical mode / chromatographic and physical conditions / special remarks; - raw data code / raw data storage place; - reduced raw data (EXPORT) file name... and all the other (regulated) information necessary. The specific info portion is always at the same position in this TEXT line and is enclosed by semicolons. This makes it possible to standardize the reporting procedure.
The structure of this EXPORT data :
time (tms) width (b05) area (%) height index 28.423 0.8855 0.09477 14.7 0 34.772 0.9221 0.04280 5.9 0 continued by about 170 or more data lines and ending with the above mentioned two information text lines: one which is made by the software transferring information already given prior the start of the analysis like: the complete date and the precise time of the injection. This time value is a very powerful identifier. If given by the second there is no uncertainty about the strict correlation of a sample, the chromatographic instrument and the responsible chromatographer. An automatic dead time calculation program corrects a non precise “injection time value” up to the 0.001 second level.
Many further information are only relevant at the time of integration, thus the finish of the EXPORT file details may also need a last text line for remarks. As the EXPORT data file is written in ASCII, it is easily readable and can be transmitted error free under the very intensive quality control modes for e-mail transmission. We never had trouble in sending and exchanging EXPORT files globally. Security exists with e-mail and ASCII attachments.
|