Line chromatograms for an optimized comparison of data with wide ranging concentrations or strongly differing numbers of substances.
Some chromatograms are really not easy to handle. Although it looks, that a simple gas analysis is easy to read, it in fact it combines a wide range of concentration with a wide or very wide range of substances of differing molecular weight. Thus isothermal analyses are either quite limited in the substance range from methane to hexane plus backflush or temperature programmed analyses with a wide range of mol weight substances are confronted with quantitative working range limitations. In order to get a correct signal for the first and largest peak = methane often in the range of 70-99 %, 100 further substances exist only in or below the 100 ppb range of concentration. Thus such chromatograms are often over sampled to make traces measurable. It causes a falsified methane signal. This also falsifies ALL other substance concentrations.
To get wide range GC chromatograms done fast enough, the separation capillary must work with a steep temperature program. This causes quickly trouble with the starting temperature because of a wide temperature range to master. The cooling down time might be too long.
How can a LOG/LOG line chromatogram comparison help ?
In the following we check line chromatogram modes to compare two chromatograms with starting temperature problems. The reason to compare was an urgent case of lost sample. The run of one sample could not be repeated. Only the two EXPORT data files existed.
First trial: see the two linear chromatograms with log time scale (this is used because of temperature programming) and a linear quantity scale. To better see traces the quantity scale sensitivity is set to 1000%, that is a factor 10 of enlargement. On the y-axis we see the quantity value of the largest peak. It is 96.5 %. With other words: the total of all other substances in the sample adds up to 3.5 %. Figure 1 shows, that the retention time values of sample “CC0000556.EXP” differ strongly from the values given with sample “CC0000843.EXP”. These are Standard EXPORT data files.
|