Quantity errors in PLC are often so large
that this excellent analytical tool is disqualified as applicable only for semi quantitative analyses. On the other hand quantitative PLC - in its full instrumentalized version - is not replaceable for very many analytical tasks.
Well done it is as precise as the best HPLC method but more important: the danger of wrong identification is very much smaller than with HPLC. As PLC users repeat sampling and separation much more often than HPLC users, the danger of excessive quantitative errors is again much smaller with PLC than with HPLC. NOTE that structure identifying highly sensitive spectroscopic detectors are applicable on-line PLC even in their micro version. The first direct PLC/MS coupling was made by the author in the late sixties but there was not yet a real demand for at these early years.
The reasons, why PLC is in quite a good shape with respect to HPLC IF WELL DONE: PLC users have for each sample a freshly unused “column”. PLC users can do multiple simultaneous runs and critically compare data. Well informed PLC users have thousand different and highly specific detection modes available and very many more mobile phases than HPLC users. PLC users can easily multiply the separation power of a plate by multi dimensional separations - two dimensional runs are only a question of proper sampling and the use of proper plates and chambers. PLC is highly sensitive and especially applicable to trace analyses.
As example: drinking water quality analysis MUST be done quantitatively, which is possible by PLC for more than hundred toxic traces simultaneously in the ppt-range of concentration (1 ppt = 1/1000 ppb; expressed as weight-concentration; 1 ppt equals 10 to the power of minus 10 weight %)
PLC users can easily and directly on-line combine chromatographic techniques with capillary GC, and HPLC/capillary HPLC. They can use HPLC as a quite helpful sample cleaning and sample preparation mode for the final PLC analysis. PLC is unbeatable for environmental mass analysis of solid and liquid “dirt” which a HPLC user would never inject into his delicate instruments. Although circular PLC is widely unknown, it is more powerful than linear PLC especially for micro preparative chromatography in its “rotation PLC mode (Sz. Nyiredy, see PLC literature)”. PLC can be used under high pressure (easily up to 3000 kg mechanical forces on a 50 x 50 or 100 x 100 mm circular plate). There is also forced flow linear PLC available with promising possibilities for mass screening analyses not only in pharmaceutical research and development. PLC is the only mode applicable in a technically underdeveloped area with missing electrical power and a minimum of space, especially to control product falsification in medicine. PLC can be done in less than 30 seconds for up to 12 samples simultaneously by micro anti circular separation and in a few minutes for up to 80 samples simultaneously. The only problem is then that our to days computers are still too slow for a complete timely quantitation of such a high data stream.
To switch from classical PLC with large mobile phase troughs and large classical plates into high performance PLC = HPTLC is a question of seconds and quite a small investment: - take HPTLC plates - replace grandfathers trough by a flat small trough or the well designed vertical chambers. - use only a few percent of the former mobile phase volume. - use the gas phase with all its analytical power for improving drastically the qualitative part of liquid-liquid or liquid solid chromatography. - or go directly to micro circular HPTLC, which to our opinion is the future of fully instrumentalized planar chromatography with direct photo quantitation.
Modern and fully instrumentalized PLC is the fastest and most economical mass analysis method chromatography has to offer.
And why are much too often PLC data so badly reproducible, that this analytically most powerful technique has in many laboratories the level of a “semi quantitative” chromatography mode often even not anymore in use ?
* The main source of quantity errors in PLC standard as well as trace analyses is the plate structure. No existing PLC stationary phase is free from structure which kills the analytically possible data quality at any wave length or reduces it by a full factor of ten.
* A next source is the fundamentally non linear calibration function for each substance in any concentration range but the use of the “rule of three” evaluation mode or by the use of data transmission concepts.
* The third source is the quite low separation power of PLC resulting in substance overlapping which often causes quantitative side effects falsifying the correct quantity value.
* The fourth source is the still missing or widely underdeveloped planar quantitation techniques, starting with digital photography at any wavelength range but not yet ending with enough accurate and enough precise PLC quantitation software.
There are quite some further error sources which are in discussion in a series of publication in the only international PLC journal “JPC, Journal of Planar Chromatography - Modern TLC”; Springer; ISSN 0933-4173. Contact the Editorial Assistant under eufeps@axelero.hu
|